Hi

First thing would be a touch test on the heat sink. It should be warm, but not to hot to touch. Heat sink down should be hotter than vertical. That assumes there's no moving air.

Next, start watching the VCXO voltage. Normal readings are around 5 to 7 volts. I'd bet you are running closer to the rails. If so a VCXO repair may help.

Past that it gets complex. The end of life on a Rb is indeed loss of lock...

Bob  



On Sep 16, 2010, at 2:51 AM, d.seiter at comcast.net wrote:

> I got one of the ebay LPRO-101s about 6 months ago and played with it for a while, but had problems maintaining a lock. First of all, it has a heat sink bolted to the bottom that is just a little larger than the LPRO itself, including being about 1.25" thick( 1" fins). Typical frequency when locked is 10,000,000.007 on a 5360A clocked by a Z3801A. 
> 
> 
> Both then and now, it takes about 45 seconds to lock from cold, and will stay locked for about 45 minutes. I found that in the position with the heat sink on the bottom will stay locked the longest (up to a few days), but then it becomes intermittent. Other positions will lock for a while, but bottom down always works the longest. 
> 
> 
> For those of you who employed these in TM5xx or similar plugins, did you have lock issues? Is my unit just old? 
> 
> 
> Dave 
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
> 



More information about the time-nuts mailing list