On 10/13/2012 11:19 PM, paul swed wrote:
> Poul
> If it is dying there should maybe be a lamp voltage you can check and see
> what it indicates. If you consider it dead then you can always checK to see
> is it something going on like low RF or maybe the RB truly is an issue. I
> have recovered some olf FRS by reheating the rb bulb and essentially
> getting the RB that collects on the bulb back into circulation.
What I did was that I heated it up with an air gun. First try, it
depleted onto the glass, but then I realized that I had left the
collection dot to the side, so it had to deplete on the colder glass,
but when I turned the collection dot to the top of the lamp and then
heated, as the hot atom raised to the top, they would deplete into the
rubidium dot and the glass surface would be free of them. The later is
important, because if they deplete on the glass, then they will absorb
the wavelengths of rubidium and this is the energy you want to get out
of the lamp.
I was fortunate as the R&S rubidium is built such that you can get the
rubidium lamp out of the clock without even turning it off. The special
tool you need is part of the assembly.
Looking at the graphs, you have had some intermittent operations with
glimpses of operation. I wonder if this is the lamp. You can open it up
and check the lamp, it may be part of your problem. I would look for
contact problems, like bad solder joint or something.