[time-nuts] homebrew counter new board test result

Tom Van Baak tvb at LeapSecond.com
Sat Feb 28 12:10:57 EST 2015


>> A good paper to read about the trouble when DUT is close (or "equal" to REF) is:
>> http://literature.cdn.keysight.com/litweb/pdf/5990-9189EN.pdf
>> "Isolating Frequency Measurement Error and Sourcing Frequency Error near the Reference Frequency Harmonics"
> 
> Nice reading and illustration of the problem. It does not go into 
> explain where these errors come from.
> 
> I especially like that he worked on an offset frequency to handle source 
> issues and that he elaborates with both time-base time and frequency 
> offset, as well as average and peak-to-peak values.
> 
> Have you seen any papers going into depth about that?

The Robert Leiby (5990-9189EN) paper was a real find. Agilent sent it to me after I ran tests of their new 53230A counter. I had two of them on loan (TCXO and OCXO) and the closer I looked the less I was impressed. The one feature that was a show-stopper for me was that the TCXO version would not outperform the OCXO version even if you gave it a BVA or maser as external reference to the counter.

That means in order to get decent performance out of the 53230A you must buy the overpriced OCXO version of the counter. I wonder if anyone else has run into this? Maybe my eval units were out of spec. Or, I wonder if anyone has opened their 53230A and hacked the timebase PLL to overcome this problem?

Anyway, that led to me checking out a pile of 53132A counters to see how well they performed.

In these tests I like to use slightly drifting, ultrastable, independent inputs instead of the old "BNC tee" trick where CH1=CH2 or CH1=CH2=REF. What you want is to see is not only the RMS noise in a measurement, but also how consistent the TI measurements are across the entire fundamental period of the inputs or timebase.

For my test I used two ultrastable sources with 1e-12 or 1e-11 frequency offset. At 1e-11 you can scan an entire 100 ns period in 10,000 seconds (under 3 hours). I'd have to look at my notes to see what I did with the REF input. I think I tried REF=CH1 and REF=CH2 and REF=3rd independent source. But the main goal was to see the interaction between CH1 and CH2 because that's the mode used by any TI measurement.

> Enrico have been looking at post-processing filtering, and it's effects.
> 
> I just haven't seen any paper giving much about how cross-talk and such 
> affects non-linearity and post-processing such as frequency reading and 
> ADEV. I have my own model from experience and various sources, but not 
> seen anything comprehensive.
> 
> Cheers,
> Magnus

Right, there's no paper that I could find yet. Instead I was planning on taking several models of popular high-end TI counters (SR620, HP5370, CNT-91, 53132A, 53230A) and run them all through the same offset/linearity test to investigate the "fine structure" of their measurement errors, as was done in the Leiby paper. If you have some measurement setup suggestions beyond what I already did with the 53132 counters, let me know.

Thanks,
/tvb