[time-nuts] R&S XSRM Rubidium Standard
KA2WEU at aol.com
KA2WEU at aol.com
Sun Sep 17 14:45:33 EDT 2017
correct !
In a message dated 9/17/2017 1:51:30 P.M. Eastern Daylight Time,
cjaysharp at gmail.com writes:
Yes.
An instrument with a calibration certificate is not necessarily accurate
but it's inaccuracies are known and can be compensated for (but only to the
accuracy of the calibration reference of course.)
On 17 Sep 2017 17:39, "Magnus Danielson" <magnus at rubidium.dyndns.org>
wrote:
> Hi,
>
> The word "calibration" is overloaded with multiple meanings, and
> incompatible too.
>
> "calibration" is often used to describe adjustments to make a device
> operate correctly, such as passing the performance checks.
>
> "calibration" in legal traceability is about measure the performance
> against references to form a traceable record of deviations from the
norml.
> This may include adjustment to ease compensation, but this is not
> necessary. Regardless of wither adjustments where done or not, the
> calibration record will indicate the errors that then needs to be applied
> to the measurement for the measurement to be traceable, and this in
itself
> requires documented knowledge about how to do the measurement.
> Otherwise it's just a fancy indication.
>
> Adjustment to a reference thus do not imply legal traceability, or even
> full functionality.
>
> For full functionality, you have to go through the performance check and
> see that all values is within limits.
>
> "calibration" can thus imply different things.
>
> I regularly see people use these terms inconsistently. That people get
> disappointed when they get the wrong thing is to be expected.
>
> Cheers,
> Magnus
>
> On 09/17/2017 05:23 PM, Scott McGrath wrote:
>
>> As to the point most modern instruments have self calibration, Most of
>> the time 'calibration' is simply the performance check adjustments are
not
>> performed unless necessary
>>
>> The difference being the instruments used in performance test are
>> traceable to a national standards body.
>>
>> So whats referred to as calibration is in reality performance
validation.
>>
>> How do I know this by becoming friendly with the local lab and years ago
>> when i worked for govt i used to moonlight at one of the local cal
labs.
>>
>> On Sep 17, 2017, at 8:57 AM, KA2WEU--- via time-nuts
<time-nuts at febo.com>
>>> wrote:
>>>
>>> Modern test and radio equipment have self calibration capabilities,
older
>>> analog do not. Calibration is not always need for just simple test,
but
>>> for specification conformation it is useful. A bit of luck also
helps.
>>>
>>>
>>>
>>>
>>> In a message dated 9/17/2017 8:08:00 A.M. Eastern Daylight Time,
>>> drkirkby at kirkbymicrowave.co.uk writes:
>>>
>>> On 15 Sep 2017 10:45, "Scott McGrath" <scmcgrath at gmail.com> wrote:
>>>>
>>>> Precisely my point, But when purchasing i expect to pay for a
>>>>
>>> calibration at a minimum.
>>>
>>> I have on occasions requested sellers to send an item to the
>>> manufacturer
>>> (Agilent or Keysight) for calibration *before* shipping it to me,
>>> offering
>>> to pay the calibration cost, but stating that I expect a full refund
if
>>> the
>>> item fails the calibration.
>>>
>>> If a test equipment dealer is confident that something is working
well,
>>> they should not object to sending it to the manufacturer for
>>> calibration,
>>> as long as the buyer is willing to pay.
>>>
>>> Of course if a seller knows little about something, they are not
going
>>> to
>>> do this, but the item should be appropriately priced.
>>>
>>> One UK seller (grace1403) declined to send an Agilent N9912A FieldFox
to
>>> Agilent, because "Agilent were too fussy"., failing items for trivual
>>> issues. But he did agree to send it to one of the cal labs he uses.
I
>>> thought it was a waste of time going to one of the less fussy outfits,
>>> but
>>> bought it anyway. It was then clear on receipt that it was faulty.
(The
>>> spectrum analyser functionality was ok, but it didn't work as a
network
>>> analyzer). He took it back, but then advertised it on eBay 6 months
>>> later. When asked, he said nothing had been done to it.
>>>
>>> eBay rules about who pays the return shipping charge for an item that
is
>>> "not as described' keep changing, and may be different on different
>>> sites.
>>> But on a heavy item shipped internationally, the postage cost can be
>>> comparable or exceed the calibration cost.
>>>
>>> Dave.
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts at febo.com
>>> To unsubscribe, go to
>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>> and follow the instructions there.
>>>
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts at febo.com
>>> To unsubscribe, go to https://www.febo.com/cgi-bin/m
>>> ailman/listinfo/time-nuts
>>> and follow the instructions there.
>>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts at febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/m
>> ailman/listinfo/time-nuts
>> and follow the instructions there.
>>
>> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/m
> ailman/listinfo/time-nuts
> and follow the instructions there.
>
_______________________________________________
time-nuts mailing list -- time-nuts at febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
More information about the time-nuts
mailing list