Randy,

I have two Fluke 731B's, one black face, one white face, both from the
'bay'.  I had to fabricate a battery pack for each of them using standard AA
or AAA cells and a collection of battery holders and double sided foam tape
to hold the packs to the 'battery bracket'.   If my memory serves, the
batteries I found were within 100 mAH capacity of what the manual called for
and everything has worked well so far.  I have never sent them for
calibration.  I use them to have some reasonable reference in my workshop.
I have two Fluke 8050A's and an HP 3478A all of which were calibrated in
their past lives but never by me.  There is agreement between the 8050A's.

At 10 Volts, one 731B reads 10.000 and the other reads 9.999 on the 8050A's.
On the 3478A, one reads 10.0003 and the other reads 10.0005.  On 1 V, one
reads 1.000 and the other reads 0.999 on the 8050A's.  On the 3478A, they
both read 1.00003.

>From this I can conclude that everything agrees within about 1 mV and,
probably, everything is 'reasonably' accurate, at least for my purposes.
The other conclusion would be that everything in my entire shop is off by
roughly the same amount.  Since I have 5 independent pieces of data, the
odds are the first conclusion is correct.

I have no resistance standards but would like to get some.  I would welcome
any suggestions.  Low maintenance (i.e. high stability) is desirable.

Joe

-----Original Message-----
From: volt-nuts-bounces at febo.com [mailto:volt-nuts-bounces at febo.com] On
Behalf Of Randy Scott
Sent: Wednesday, December 30, 2009 10:54 AM
To: volt-nuts at febo.com
Subject: [volt-nuts] Poll


Nuts,

I'm curious to know if people are maintaining their own standards of voltage
and resistance.  If so, do you have a single standard or multiple?  How
often do you send your standard(s) for calibration?  How often do you
perform comparisons between your multiple standards?

Do you maintain a standard "just because" or do you use it to calibrate
other equipment (e.g., artifact calibration of a multimeter)?

Thanks.

Randy.


      