[volt-nuts] Which meter?
[email protected]
the-al-bundy at home.nl
Sat Nov 14 20:56:45 UTC 2009
Hi Greg,
Well I've made a design with the LTZ1000A and must test it's stability. The
best DVM we have is the 34401, but with the 6.5dig you don't get far with a
LTZ1000 as DUT. So we need something better. As far as I've seen the only
options are the 8508A and the 3458A.
The 8508A is relativly new (2002 I thought), so one would expect better
specifications than the 3485A and more options like other connectivity than
GPIB.
However the 3458A is already around there for >20years and it seems to be
one of the best DVM. The 100k sample rate is a feature which could be handy
in the future.
Money is not so much an issue, just we have to make a good investment for
the next 5+ years or so.
Of course we expect to use the DVM more often for ultra stabile
circuits/sources (in currents as wel as volts).
Thanks,
Bram
----- Original Message -----
From: "Greg Burnett" <gbusg at comcast.net>
To: "Discussion of precise voltage measurement" <volt-nuts at febo.com>
Sent: Saturday, November 14, 2009 6:52 PM
Subject: Re: [volt-nuts] Which meter?
> Hi Bram,
> If you want to make high-speed measurements, the Agilent 3458A is the
> clear
> choice (because it can take up to 100,000 readings per second). The same
> reasoning would apply if you need to use your DVM to make precision
> digitization of waveforms. (For example the 3458A's triggered, digitized
> measurements can be used to evaluate the linearity of a triangle
> waveform.)
>
> For basic DC linearity (e.g., from full-scale to tenth scale or lower),
> the
> 3458A is as good as it gets, as far as I know.
>
> Having said that, there might be a particular application where the Fluke
> 8508A would be your best fit? But I'm not sure about that. Do you have
> specific measurement applications in mind?
>
> I've personally always been extremely impressed with the 3458A
> performance,
> so unless I had a specific measurement application (that I'd researched
> and
> found that was addressed better by the 8508A), I'd buy the 3458A.
>
> Best,
> Greg
>
> ----- Original Message -----
> From: "The-al-bundy at home.nl" <the-al-bundy at home.nl>
> To: <volt-nuts at febo.com>
> Sent: Saturday, November 14, 2009 8:03 AM
> Subject: [volt-nuts] Which meter?
>
>
> Hi,
>
> I'm at a point to purchase a new 8.5dig DVM. Choices are the (good old HP)
> Agilent 3485A or the Fluke 8508A. The Fluke is almost 2x as expencive...
>
> My feelings say that the 3485A is still the best one out there, is that
> true?
>
> Thanks,
> Bram
>
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
More information about the volt-nuts
mailing list