[volt-nuts] Schematic error in Solartron 7150 Plus service manual
David C. Partridge
david.partridge at perdrix.co.uk
Wed Nov 10 17:34:39 UTC 2010
An updated version of the manual with schematic corrections has been provided by Nigel Clarke (GandalfG8 at aol.com) and has been uploaded to Didier Juge's website <www.ko4bb.com>
Dave
-----Original Message-----
From: volt-nuts-bounces at febo.com [mailto:volt-nuts-bounces at febo.com] On Behalf Of David C. Partridge
Sent: 09 November 2010 10:59
To: volt-nuts at febo.com; TekScopes at yahoogroups.com; TekScopes2 at yahoogroups.com; hp_agilent_equipment at yahoogroups.com
Subject: [volt-nuts] Schematic error in Solartron 7150 Plus service manual
If you ever need to work on the Ohms circuit of a Solartron 7150 Plus, be warned, there is a schematic error in the service manual. In diagram 2.1 (PCB1, Floating Analog), the top of resistor R22 should connect to the Source terminals of TR4, TR9 and the top of R26. This "wire" should NOT connect to the gate of TR4 as it appears to do in the schematic.
I couldn't work out what was going on here for a while, and then the penny dropped!
The schematic in the original 7150 Service manual is correct.
Regards,
David Partridge
_______________________________________________
volt-nuts mailing list -- volt-nuts at febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
More information about the volt-nuts
mailing list