[volt-nuts] A 7081 with a MAINS FAULT problem
David C. Partridge
david.partridge at perdrix.co.uk
Sat Apr 21 00:23:57 UTC 2012
Follow up - if I set S1 on the Earthy logic board so that the switch labelled "RESET OFF" is to the left (ON I think), then I just get the MAINS FAULT message displayed solidly. This would suggest the earthy processor stops polling $4C00 when it detects a MAINS FAULT condition, with the result that it will go into reset as the counter (IC5) isn't being reset.
I initially found the schematic for the earthy logic board a little confusing as it appeared that this switch position was only connected to a pullup resistor in IC12 (pin 7) which it grounded when the switch was made. It took me a while to spot this was also connected to IC6 pin 2!
So the fault would seem to be on the floating board. Pin 19 of the floating CPU is being fed with a nice 50Hz square wave.
I'll try swapping the earthy boards again - just to be sure ...
Dave
-----Original Message-----
From: volt-nuts-bounces at febo.com [mailto:volt-nuts-bounces at febo.com] On Behalf Of David C. Partridge
Sent: 20 April 2012 20:24
To: 'Discussion of precise voltage measurement'
Subject: [volt-nuts] A 7081 with a MAINS FAULT problem
This is the unit with the dead VFD. I've borrowed the front panel and VFD from a 7071 to allow me to try to fix the other faults.
Initial symptoms MAINS FAULT followed by WAIT @F.
After replacing the blown +15V regulator on the floating board, all that I got was a MAINS FAULT message which cleared and was followed by a WAIT @F.
I tried substituting the earthy logic and earthy CPU boards from a known good meter, and that made no difference. The link cable between the two boards tests good.
Replacing the floating CPU (IC801) with the one from the good meter results in a MAINS FAULT message being flashed at me at about 3 second intervals. Sometimes I see a WAIT @F between the MAINS FAULT messages.
As far as I can tell, the ROM (IC803) and RAM (IC826) make no difference to the behaviour.
Scoping pin 12 (floating to earthy) of the floating CPU I see what may be one character being sent back to the earthy board at about the time the MAINS FAULT message clears. (-15V for 28uS, -10V for 28us, -15V for 36us, then back to -10V).
Scoping the emitter of TR1 on the earthy board (or pin 11 of the floating CPU) appears only to show one 120uS long low pulse at about the same time.
I'm not seeing any obvious DUMP NVM command being sent unless that single low pulse is the DUMP NVM command, and there's no sign of a reply.
Regards,
David Partridge
_______________________________________________
volt-nuts mailing list -- volt-nuts at febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
More information about the volt-nuts
mailing list