[volt-nuts] Thermistor value for 732A Reference 10V Standard
DaveH
info at blackmountainforge.com
Fri Jun 29 04:02:46 UTC 2012
Are you reading this in-circuit and operating? The current flowing through
the resistor will skew the reading.
Stupid question I know but had to be asked...
DaveH
> -----Original Message-----
> From: volt-nuts-bounces at febo.com
> [mailto:volt-nuts-bounces at febo.com] On Behalf Of Mitch Van Ochten
> Sent: Thursday, June 28, 2012 18:43
> To: Discussion of precise voltage measurement
> Subject: Re: [volt-nuts] Thermistor value for 732A Reference
> 10V Standard
>
> Hi Tom,
>
> According to the Fluke 732A manual page 2-6 the nominal value
> is between
> 3k - 4k with the oven at normal operating temperature. Mine
> checks 3.545 k
> and has been within a few ohms of that reading each year for
> five years. You
> may have a heater problem. There is a protective thermal
> fuse which can go
> open, in series with the heating elements. It's purpose is
> to prevent a
> runaway oven from cooking everything. Once it opens you will
> never get any
> heating until it is replaced.
>
>
> Regards,
>
> mitch
>
>
>
> ----- Original Message -----
> From: <Leedyt at aol.com>
> To: <volt-nuts at febo.com>
> Sent: Thursday, June 28, 2012 7:04 PM
> Subject: [volt-nuts] Thermistor value for 732A Reference 10V Standard
>
>
> >
> >
> > Hi!
> >
> > Can anyone give me a rough value of the thermistor
> resistance that is
> > brought out to the front panel of a Fluke 732A DC Reference
> Standard
> > (lower
> > left-hand side)? The thermistor is for the user to monitor the
> > temperature of
> > the oven that contains the voltage reference. I see a
> resistance that
> > starts out at 6600 Ohms (unit was nearly cold) and climbs
> and levels out
> > at
> > about 8400 Ohms after a day, or so. Is this reasonable?
> The thermistor
> > is
> > given in the parts list as RT2, a Fenwal JA41J1 (no longer
> made) and on
> > the
> > schematic as a 10K @ 25C unit. So this doesn't make sense.
> >
> > Does anyone have any experience with this? Or is just
> simpler to replace
> > the thermistor with a 100 Ohm RTD and be done with it? Or
> would this
> > drive
> > the calibration labs nuts?
> >
> > Any references or advice would be appreciated.
> >
> > Tom Leedy
> > Clarksburg, MD
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts at febo.com
> > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> > and follow the instructions there.
>
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
More information about the volt-nuts
mailing list