[volt-nuts] Fluke 8840A 100 and 1000 VDC ranges read low.

Rob Klein rob.klein at smalldesign.nl
Tue Mar 13 15:27:33 UTC 2012


John,

I do not know what these "instructions" you have are, but the 8840A manual
(http://assets.fluke.com/manuals/8840A___imeng0300.pdf), pp. 112 - 114, 
describes
the calibration procedure in some detail. And yes, that includes the 
200V and 1000V
ranges.

- Rob.

Op 12-3-2012 23:13, John Phillips schreef:
> Hi,
> I have calibrated several Fluke 8840A DMM but have never seen one that
> needed the 100 or the 1000 VDC range calibrated. The instructions I have do
> not say anything about calibration constants for these ranges.
> I have changed the 100X voltage devider but that did not help.
> 10 VDC reads 9.9997
> 10 VDC on the 100 scale reads 9.985
> 100 VDC reads 99.915 on 100 scale and 99.81 on the 1000 scale.
> 1 VDC on the 10 scale reads 0.9999
> The 100 to 1 dividers are about 99.995 and 99.997 to one. The 99.997
> divider is in the unit now.
>
> What did I not do to get this to work. There must be a step someplace to do
> a math cal on the 100 to 1 divider.
> Thanks you...
> .