[volt-nuts] Fluke 731A Transfer Standard and Fluke WireWoundResistors
Roy Phillips
phill.r1 at btinternet.com
Tue Sep 18 08:56:38 UTC 2012
Joe
I guess that there are many of us with a reference board from the HP3458A
who would be pleased to have a good design making use of this item. For
those of you who have been around a long time, I recently obtained a
Weston - Rotex DC Absolute Voltage/Current Standard which came from a
Standards Lab. Much to my surprise it still performs very well, and its
interesting to note that each standard output is delivered to a separate
pair of terminals, unlike the Fluke 731B, but more in keeping with the
latter models. So it would seem that the old-timers knew a thing or two
about 'high standards', but at a price of course. It required a mass of
components to produce the result - it would appear to be 1960's
design/manufacture. Are there many of these out there ? - I would be
interested to see its specification as I cannot find any reference to this
item.
Roy
-----Original Message-----
From: Fred Schneider
Sent: Monday, September 17, 2012 3:29 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Fluke 731A Transfer Standard and Fluke
WireWoundResistors
My 731A also drifts all over the place. The LM399 I am making was first
ment to become a replacement for use in the 731A, and like you, use the
diverder. But not any more, I have a LT1027B reference that is on 24/7 for
about a year and nice behaving. The 1027 behaves better as my 731A and
delivers 10V . Weak point is the resitors i used to adjust it to 10V. ( I
was young and.... ;-) ) I plan to change that using better ones and a better
construction and after that, use that to feed the divider of the 731A. Also
replaced the opamp of the 731A for a better one but that improved only a
little.
Then use a chopper opamp with LM10 as buffer on the output of the 731A so
its output impedance will be lower and it will be able to be used direct
without a null detector, or before a KV devider without collapsing.
Fred PA4TIM
Op 17 sep. 2012 om 14:38 heeft "J. L. Trantham" <jltran at att.net> het
volgende geschreven:
> Pete,
>
> As far as the connection goes, the current resistor wire appears to be
> 'soldered' to what appear to be gold plated steel wires (very stiff yet
> bendable). If the temperature is relatively stable, it should not be too
> bad a problem as far as the repair goes. However, it seems to drift all
> over the place.
>
> I changed the Op Amp to no benefit.
>
> I am leaning to switching to a different voltage reference. I have a
> 'spare' reference board from an HP 3458A and it would make a good project
> to
> install this, change the wiring to the Op Amp to adjust the 'other side'
> of
> the Op Amp input to adjust to 10.000000 VDC and still be able to use the
> output divider in the 731A to get the rest of the voltages offered by the
> 731A.
>
> Joe
>
> -----Original Message-----
> From: volt-nuts-bounces at febo.com [mailto:volt-nuts-bounces at febo.com] On
> Behalf Of Pete Lancashire
> Sent: Sunday, September 16, 2012 4:03 PM
> To: Discussion of precise voltage measurement
> Subject: Re: [volt-nuts] Fluke 731A Transfer Standard and Fluke Wire
> WoundResistors
>
>
> There is a good chance the wire is Manganin or an allow that is similar.
>
> If what you use to connect to the resistance wire is not the same you run
> into things like EMF (think thermocouple) being created at the repair.
>
> -pete
>
> On Sun, Sep 16, 2012 at 1:36 PM, J. L. Trantham <jltran at att.net> wrote:
>> I have been following the discussion about the LM399 heater, noting
>> mention of the Fluke Wire Wound resistors used often in Fluke
>> precision equipment.
>>
>>
>>
>> I have a 731A that I have been resuscitating and have found that the
>> R8A R8B wire wound resistor is failing. Initially, the source was
>> very unstable then went to about 8 VDC. This led me to an open R8B.
>> I was able to ‘reconnect’ the broken ends, loosing about half a turn
>> of wire. Then, again, unstable and went to about 13 VDC. This led me
>> to an open R8A. Again, able to reconnect, and, again, loosing about ½
>> turn.
>>
>>
>>
>> It appears that the wire is corroded with corrosion leading to
>> failure. This brings up several questions.
>>
>>
>>
>> 1. What are the benefits of these wire wound resistors? Tempco?
>> Ability to construct precise resistance?
>> 2. Is there a source of the resistance wire used in these resistors
> so
>> that the resistor can be reconstructed? I think the resistance is in
>> the 30 to 40 ohms per foot range. Something like 36 to 38 gauge
>> Nichrome 60 would work but it has to be insulated.
>> 3. Would it be better to look for a collection of commercial
> resistors
>> to ‘replace’ R8A, R8B? If so, what? The manual states that IC2, the
>> voltage reference, R6, R7, R9, and R10, are ‘factory selected’ and all
>> appear to be OK.
>> 4. Would it be better to completely abandon the voltage reference
>> section of the 731A in favor of a ‘new’ (perhaps LM399, LTZ1000A,
>> etc.,
>> based) reference to generate 10 VDC and still use the output divider of
> the
>> 731A to generate the remaining voltages?
>>
>>
>>
>> The 731A manual and schematic are here:
>>
>>
>>
>> http://www.ko4bb.com/Manuals/Fluke/Fluke_731A_Transfer_Standard_Instru
>> ction_
>> Manual.pdf
>>
>>
>>
>> R8A is the ‘high’ side of a voltage divider with R8B, R9, R10 and R11
>> (a 10 ohm, 20 turn pot) forming the ‘low’ side, dividing 10 VDC to
>> feed the voltage reference, IC2, a DH80417B. The op amp, IC1, used to
>> generate the 10 VDC is an LM301AH.
>>
>>
>>
>> R8A measures about 5903 ohms, R8B measures about 11558 ohms, R9
>> measures about 78.02 ohms, and R10 measures about 399.8 ohms.
>>
>>
>>
>> Thanks in advance for any and all information and suggestions.
>>
>>
>>
>> Joe
>>
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts at febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
_______________________________________________
volt-nuts mailing list -- volt-nuts at febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
More information about the volt-nuts
mailing list