[volt-nuts] HP3458a available (was Re: hp 3458a advice)
John Phillips
john.phillips0 at gmail.com
Thu Feb 7 23:10:07 EST 2013
I read most of the thread. This could be a pre-production unit. The errors
indicate it could need both the in guard and out guard CPUs. The in guard
unit (03458-05503) I think is involved with the A/D conversion and that is
where the flatness error is most likly coming from.
On Thu, Feb 7, 2013 at 7:48 PM, John Phillips <john.phillips0 at gmail.com>wrote:
> That is what I get for jumping in without reading the thread.
> I have sent 2 or 3 units in for them to repair...
> When they quit selling part I was forced to junk or have them repair.
> What parts are needed?
>
>
> on Thu, Feb 7, 2013 at 7:12 PM, Tom Miller <tmiller11147 at verizon.net>wrte:
>
>> They might not take it in without a serial number.
>>
>> Regards
>>
>> ----- Original Message ----- From: "John Phillips" <
>> john.phillips0 at gmail.com>
>> To: "Discussion of precise voltage measurement" <volt-nuts at febo.com>
>> Sent: Thursday, February 07, 2013 9:34 PM
>>
>> Subject: Re: [volt-nuts] HP3458a available (was Re: hp 3458a advice)
>>
>>
>> You should be able to any unit repaired by Agilent as long as all the
>>> parts
>>> are there for the $3000 flat rate repair and calibration charge.
>>>
>>> On Thu, Feb 7, 2013 at 5:59 PM, J. L. Trantham <jltran at att.net> wrote:
>>>
>>> Paul,
>>>>
>>>> The 'value' of a 3458A is based on several issues, not the least of
>>>> which
>>>> is
>>>> supply and demand. You only have to look at theBay to get an idea of
>>>> what
>>>> units are selling for, both working and non-working. The fact that the
>>>> unit
>>>> passes it's 'Self Test' is not a guarantee that it would be able to be
>>>> calibrated to published performance specs.
>>>>
>>>> Tha Agilent website is very useful in determining an 'upper bound' on
>>>> the
>>>> value. The 3458A is currently orderable and currently supported. A new
>>>> 3458A (without options) is available at $9,085 and would come with a
>>>> fresh
>>>> calibration and warranty.
>>>>
>>>> Repair Per Incident is available for $2,660.64. This means that if you
>>>> have
>>>> a complete unit, they will repair it for that price. That also comes
>>>> with
>>>> a
>>>> fresh calibration. I don't know if that price includes all recommended
>>>> upgrades as listed in the Service Notes, of which there are 20, 6 of
>>>> which
>>>> list 'Modification Recommended' and one of which lists 'Modification
>>>> Available'. Perhaps someone on the forum knows the answer to this.
>>>> Certainly Gary Biermann at Agilent can answer this.
>>>>
>>>> The fact that your unit does not have a serial number sticker but
>>>> instead
>>>> 'MTG PROTO 27' would be concerning as to whether Agilent would consider
>>>> this
>>>> a 'complete 3458A', and thus a candidate for repair, or, rather, a
>>>> 'prototype' or some other descriptor that would imply that it looks
>>>> like a
>>>> 3458A but is really not a 3458A.
>>>>
>>>> Both new and newly calibrated units are candidates to add a Repair
>>>> Agreement, available for $178.68 per year with up to 5 year blocks
>>>> available
>>>> for purchase.
>>>>
>>>> I note you mentioned firmware REV 2,1. The current firmware is 9,1.
>>>> Therefore it would appear that you have a rather old and probably not
>>>> 'updated' unit. That plus the 'stored in an attic' history would be
>>>> concerning about the ability to repair the unit without changing several
>>>> assemblies, etc.
>>>>
>>>> Assuming you could get the unit repaired and updated by Agilent,
>>>> complete
>>>> with a fresh calibration, then the maximum value would be $9,085 minus
>>>> $2,660.64 minus what ever value you place on 'new' over 'old'.
>>>>
>>>> Then you are left with an older unit, made in the USA, with no (or
>>>> perhaps
>>>> a
>>>> reduced) warranty as opposed to a new unit, made in Malaysia, with a
>>>> warranty. The question of whether a 'new' Malaysian unit or an 'old'
>>>> American unit is better is the subject of further discussion and
>>>> speculation.
>>>>
>>>> If the unit cannot be repaired by Agilent, then, in my opinion, the
>>>> value
>>>> goes down and approaches the value of the hardware, transformer, etc.,
>>>> with
>>>> serious questions about the state and reparability of the assemblies,
>>>> particularly given the 'stored in an attic' history.
>>>>
>>>> Gary Biermann would be very helpful in getting the history and
>>>> reparability
>>>> of the unit. He would be worth contacting about these issues. He
>>>> works at
>>>> the Loveland Calibration Facility.
>>>>
>>>> Hope this helps.
>>>>
>>>> Joe
>>>>
>>>> -----Original Message-----
>>>> From: volt-nuts-bounces at febo.com [mailto:volt-nuts-bounces@**febo.com<volt-nuts-bounces at febo.com>]
>>>> On
>>>> Behalf Of Paul Fox
>>>> Sent: Thursday, February 07, 2013 8:54 AM
>>>> To: Discussion of precise voltage measurement
>>>> Subject: Re: [volt-nuts] HP3458a available (was Re: hp 3458a advice)
>>>>
>>>> part of me hates to do this, but... there have been some rule-breakers
>>>> that ignored the "under $350" limit, and now they've ruined it for
>>>> everyone. ;-)
>>>>
>>>> clearly the research i did into prices for "HP3458a, needs repair" was
>>>> woefully inadequate.
>>>>
>>>> i guess the easiest thing to do is to simply remove the "under $350"
>>>> part. that way everyone's working under the same assumptions. again,
>>>> sorry. local buyers still get preference. (and, if you'd like to chew
>>>> me out for being a bozo, that's fine, but please do it privately. maybe
>>>> i should have used ebay in the first place. lesson learned.)
>>>>
>>>> paul
>>>>
>>>> i wrote:
>>>> > resurrecting a somewhat older thread --
>>>> >
>>>> > it's now clear to me that i have little interest in (or time for)
>>>> > resurrecting my 3458a. several people on the list were interested in
>>>> > it when i first brought it up, and you were all very helpful. i'd
>>>> > rather not go through the ebay hassle, and i'd rather make it
>>>> > available to one of you folks first in any case.
>>>> >
>>>> > i've put up a small set of pictures, along with a short video of
>>>> > the startup sequence at http://www.foxharp.net/hp3458a . re-reading
>>>> > my original list message (below), i see that my startup video didn't
>>>> > capture an error message i saw back in october. so in addition to
>>>> the
>>>> > "110, Calibration required -- ACAL" message in the video, i guess we
>>>> > should assume that it sometimes gives "204, "Hardware Failure --
>>>> Flatness
>>>> > DAC Convergence: 198"
>>>> >
>>>> > i haven't opened the unit, and wasn't really planning on it. if
>>>> > there's something that you'd like to know that will be clear from
>>>> just
>>>> > looking, and if it might appreciably change the unit's value for you,
>>>> > let me know and i'll find a screwdriver and take some more pictures.
>>>> >
>>>> > i have two original (i believe) copies of the manual. no service
>>>> manual.
>>>> > firmware version is "REV 2,1".
>>>> >
>>>> > so how about this: best offer (by private mail) under $350 gets it,
>>>> > shipping not included (it looks like shipping will be in the $40 to
>>>> $80
>>>> > range, once i find a box). for obvious reasons, a local-to-boston
>>>> > transaction will be given a lot of preference.
>>>> >
>>>> > i'll wait until next wednesday evening (i.e., about a week -- feb 13)
>>>> > before choosing a lucky recipient, since i hear some people don't
>>>> read
>>>> > their mail as continuously as i do.
>>>> >
>>>> > paul
>>>> >
>>>> > i wrote:
>>>> > > thanks joe, and to the other couple of people who responded
>>>> privately.
>>>>
>>>> > > i need to figure out how much of a project i want this to be, and
>>>> how
>>>> > > it ranks against all the other projects that i also don't have
>>>> time
>>>> > > for. :-)
>>>> > >
>>>> > > i have the full users manual (two copies!), but i don't have
>>>> anything
>>>> > > else that speaks GPIB. i'll play with some of the front panel
>>>> > > commands you've suggested, while i think about what to do.
>>>> thanks!
>>>> > >
>>>> > > paul
>>>> > >
>>>> > > j. l. trantham wrote:
>>>> > > > Paul,
>>>> > > >
>>>> > > > First, the meter is worth something if nothing other than
>>>> parts.
>>>> It is
>>>> > ...
>>>> > > > Good luck.
>>>> > > >
>>>> > > > Joe
>>>> > > >
>>>> > > > -----Original Message-----
>>>> > > > From: volt-nuts-bounces at febo.com
>>>> [mailto:volt-nuts-bounces@**febo.com <volt-nuts-bounces at febo.com>] On
>>>> > > > Behalf Of Paul Fox
>>>> > > > Sent: Monday, October 15, 2012 9:50 AM
>>>> > > > To: Discussion of precise voltage measurement
>>>> > > > Subject: [volt-nuts] hp 3458a advice
>>>> > > >
>>>> > > > hi --
>>>> > > >
>>>> > > > i'm not a volt-nuts regular, and not even a precision
>>>> measurement
>>>> nut
>>>> > > > at all, but given past discussion in the list archives, i think
>>>> you
>>>> > > > folks might be able to help.
>>>> > > >
>>>> > > > i was recently given an HP 3458a multimeter. the donor
>>>> thought it
>>>> > > > worked when he last used it, but that was quite a long time
>>>> ago,
>>>> and i
>>>> > > > think it was stored in an attic for some time, so i'm not too
>>>> > > > surprised that it now doesn't pass self tests. :-/ he mainly
>>>> wanted
>>>> > > > it out of his house, along with a bunch of other vintage (sadly
>>>> mostly
>>>> > > > non-working) equipment he'd accumulated over the years.
>>>> > > >
>>>> > > > the self-test error is:
>>>> > > > ERRSTR 204, "Hardware Failure -- Flatness DAC Convergence:
>>>> 198"
>>>> > > >
>>>> > > > judging from past list traffic, repairs on these units are
>>>> > > > expensive, and it would still need recalibration to be useful,
>>>> and
>>>> > > > there are may be other parts (eeproms?) that may not be good
>>>> > > > any more. i'm not really interested in pursuing any of this
>>>> > > > myself.
>>>> > > >
>>>> > > > what's worse is that the unit doesn't have a serial number. (i
>>>> assume
>>>> > > > the s/n would normally be obvious on the rear of the unit.) it
>>>> bears a
>>>> > > > sticker on the back that says "MTG PROTO 27". i think the unit
>>>> came
>>>> > > > from a place that themselves produced reference standards, and
>>>> they
>>>> > > > may have gotten a very early model from HP to work with.
>>>> that's
>>>> > > > conjecture. i haven't opened the unit, so i don't know if
>>>> there
>>>> are
>>>> > > > other clues to its age inside.
>>>> > > >
>>>> > > > so: is this machine useful to anyone? given the error message
>>>> above,
>>>> > > > are there any measurements at all that it might still get
>>>> correct,
>>>> in
>>>> > > > it's current state?
>>>> > > >
>>>> > > > paul
>>>> > > > =---------------------
>>>> > > > paul fox, pgf at foxharp.boston.ma.us (arlington, ma, where it's
>>>> 63.1
>>>> degrees)
>>>> > > >
>>>> >
>>>> > =---------------------
>>>> > paul fox, pgf at foxharp.boston.ma.us (arlington, ma, where it's 30.0
>>>> degrees)
>>>> > ______________________________**_________________
>>>> > volt-nuts mailing list -- volt-nuts at febo.com
>>>> > To unsubscribe, go to
>>>> https://www.febo.com/cgi-bin/**mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts>
>>>> > and follow the instructions there.
>>>>
>>>> =---------------------
>>>> paul fox, pgf at foxharp.boston.ma.us (arlington, ma, where it's 19.2
>>>> degrees)
>>>> ______________________________**_________________
>>>> volt-nuts mailing list -- volt-nuts at febo.com
>>>> To unsubscribe, go to
>>>> https://www.febo.com/cgi-bin/**mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts>
>>>> and follow the instructions there.
>>>>
>>>> ______________________________**_________________
>>>> volt-nuts mailing list -- volt-nuts at febo.com
>>>> To unsubscribe, go to
>>>> https://www.febo.com/cgi-bin/**mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts>
>>>> and follow the instructions there.
>>>>
>>>>
>>>
>>>
>>> --
>>> John Phillips
>>> ______________________________**_________________
>>> volt-nuts mailing list -- volt-nuts at febo.com
>>> To unsubscribe, go to https://www.febo.com/cgi-bin/**
>>> mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts>
>>> and follow the instructions there.
>>>
>>>
>> ______________________________**_________________
>> volt-nuts mailing list -- volt-nuts at febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/**
>> mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts>
>> and follow the instructions there.
>>
>
>
>
> --
> John Phillips
--
John Phillips
More information about the volt-nuts
mailing list