[volt-nuts] Datron 1281 NVRAM woes
acbern at gmx.de
acbern at gmx.de
Tue Apr 12 23:03:26 EDT 2016
What I would think is the case here (had that on a 4920 ac standard, similar/same CPU board) is that the SRAM has failed. Replacing it means a new calibration though.
So you may wish to carefully check before. Are you in Germany? Can also drop me a PM.
> Gesendet: Dienstag, 12. April 2016 um 09:20 Uhr
> Von: "Bert K" <bkachler at hotmail.com>
> An: volt-nuts at febo.com
> Betreff: [volt-nuts] Datron 1281 NVRAM woes
>
> Hello,
>
> all of a sudden my beloved Datron 1281 stopped working.
> If briefly displays the familiar
> "DCV: Auto 100mV......"
> but then
> "Secondary NV C/SUM ERROR 2101"
> and then
> "9005.7 No Rtx 001B42"
>
> The voltages from the DC board PS seem okay, also the
> battery displays only a bit less than 3.7Vol.
>
> Obviously I wonder whether anything can be done here.
> In particular I wonder about the NVRAM setup,
> namely whether "Secondary" could be just a safety copy
> of the "Primary". Or whether this is the less important
> set of self-cal constants, etc.
> I couldn't find any info in the various manuals.
>
>
> Also the second error seems to have to do with
> serial comm, and it is not evident that is has anything
> to do with with the first error. If not, then they could have
> a common underlying reason, like a faulty PS.
>
> The PS of the digital board is different than as the one
> described in the manual, so I can't really check the voltages
> meaningfully. Does anybode have schematics of this
> second (I hear common) type of digital board?
> Or knows otherwise what the voltages on the various
> test points are supposed to be?
>
> Thanks,
> Bert
>
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
More information about the volt-nuts
mailing list