[volt-nuts] Solartron 7081 randomly re-initialises
BIll Ezell
wje at quackers.net
Thu Apr 14 12:31:48 EDT 2016
Repost from my response in the new Solartron group, :
>I assume you've eliminated obvious things like power-line spikes? One
other thing comes to mind, the power-up reset logic for the CPU, the
RESET >signal. My schematics are so bad it's hard to tell for sure, but
it looks like that's generated by R6 and C2, along with IC4 on the
earthy board. I'd check >C2, if it's very leaky, could be pulling IC4-1
down enough to give random triggering. IC6 is a possibility also, IC6-6
is used to also trigger a reset by >pulling C2 to ground.
I'd think if there was enough ripple on the 5V supply to trigger a
reset, you'd be seeing other bad things.
Regading the nvram, the comments on that are dead-on, my cross post
about that:
>The cal constants are in a non-volatile ram, it's possible you'll need
to replace that. I've never seen a failure, but I've heard anecdotally
that they can >fail. It's also possible that your meter has been sitting
unused for so long that the nv-ness of the ram has leaked away.
>They work by injecting a stored charge into effectively a
well-insulated gate. Over many years, the charge can leak away. That's
why there's a rewrite >cycle the meters can do.
I forget what the nvram refresh command is, but it's in the manual.
Also, the nvram keeps two separate copies. I thought it would
automatically try to read the second copy if the first was bad, and
automatically rewrite the bad one. But, if enough charge has leaked,
there could be multiple errors across the entire nvram.
More information about the volt-nuts