At least in my experience, such documents do not exist. What happens:
1) The unit is put on a calibration schedule (1 year) on an arbitrary basis
2) The tech is expected to read the manual for the device and calibrate it
3) Any deviation from cal is noted in generic documents
Put another way, a non-Cesium standard is no different than a DVM. If it’s a
Cesium, it’s a primary standard. There is no calibration.
Bob
> On Mar 11, 2016, at 1:13 AM, timeok at timeok.it wrote:
>
> Hi all,
>
> Among the many experts and documentation available on the internet my question might seem trivial .
>
>
> I'm looking for a written document , not generic information, describing the procedure , timing and methods of calibration of one or more standard present in a laboratory frequency standard . This document should also describe the process of selection of the times frame between two successive calibrations and other parameters that have been decided according to the type of frequency standard to be calibrated respect the target and so on .
>
> Imagine that a new employee receives the manual of procedures to maintain calibrated some frequency standards such as a Caesium , a Rubidium an OCXO and a GPS in a standard's room.
>
> I know, and I have, several documents and book describe all the generic procedure but, my question is if is it available a complete document ready to use about these monthly operations.
>
> thank you,
>
> Luciano
> www.timeok.it
> Message sent via Atmail Open - http://atmail.org/
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
More information about the time-nuts
mailing list