[volt-nuts] vSolartron7081 rom reading
m k
m1k3k1 at hotmail.com
Tue Jan 15 22:29:30 UTC 2013
I just noticed a gotcha, I need to set the key to calibrate before I may issue those commands, I guess I need a key. I have printed off the photo and measurements someone on this board did, now to get to a key shop...
M K
> From: david.partridge at perdrix.co.uk
> To: volt-nuts at febo.com
> Date: Sun, 13 Jan 2013 17:14:07 +0000
> Subject: Re: [volt-nuts] vSolartron7081 rom reading
>
> If all you want is to save the calibration factors, just issue CALIBRATE,DUMP command
>
> To refresh the EAROM calibration factors, issue CALIBRATE,REFRESH - barring component failure that should be good for another ten years or so. You can also display the NVM contents with the MONITOR command, but NOT re-write them.
>
> The only way that I am aware of that will write the caibration constants if they have been lost is to run a full calibration.
>
> Again all this in chapter 8.
>
> I don't know which ROM programmers that can read or write the ER3400 (just ignorance - there may be loads).
>
>
> Regards,
> David Partridge
> -----Original Message-----
> From: volt-nuts-bounces at febo.com [mailto:volt-nuts-bounces at febo.com] On Behalf Of m k
> Sent: 13 January 2013 16:57
> To: volt-nuts at febo.com
> Subject: Re: [volt-nuts] vSolartron7081 rom reading
>
>
> Hi David,
>
> well it is currently working and close to calibrated as far as I can tell. But I want to get those calibration factors saved before it dies on me. Last calibrated 2008, and well down on drift by now.
>
> I get a diurnal 0.14 microvolt drift with a shorting plug that stays on when not in use. a drift of about 0.1 microvolts in about 3 weeks. Nothing else to confirm anything else as I have only just got it chatting out through the rs232.
>
> I will have to buy something to read them with.
>
> M K
>
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
More information about the volt-nuts
mailing list